In very long baseline interferometry, it is typical to use a hydrogen maser atomic clock to sample data. However, given how expensive these atomic clocks are, it seems surprising that more observatories aren't using non-local atomic clocks, such as GPS clocks.

This Wikipedia article says "The antenna signal is sampled with an extremely precise and stable atomic clock (usually a hydrogen maser) that is additionally locked onto a GPS time standard." My question is, why not do away with the hydrogen maser entirely?

$\endgroup$
2
  • 1
    $\begingroup$ I'm not qualified to answer, but I can venture a guess. In VLBI such as that discussed in this question the wavelength is only 1.3 mm, ~230 GHz, you need sub-picosecond short term stability, and I don't think GPS offers anything even close to that. $\endgroup$
    – uhoh
    Dec 9, 2018 at 7:02
  • 2
    $\begingroup$ GPS offers ca. 40ns precision, which doesn't seem to be enough. $\endgroup$ Dec 12, 2018 at 15:18

1 Answer 1

7
$\begingroup$

There is a presentation that covers this subject which is a regular at the International VLBI Service workshops. The most recent version is at High-accuracy Time and Frequency in VLBI and the relevant derivation of needed accuracy is on slide 5. Basically for interferometery, you are trying to match up the phase of the waveforms (position along a sine wave in the simplistic sense) of the two signals from each of the dishes.

If we use an observing frequency of 10 GHz (10e9 Hz) and want to keep the phase coherent to ~10 degrees (out of the 360 degrees) after 1000 seconds integration then you need 10 / (360 * 10 * 1e9 * 1e3) or 10 / 3.6e16 or a clock stability of about 2.8e-15. Higher frequencies such as the 230 GHz mentioned in the comment will be even more demanding.

If you look at Allan Deviation/Variance plots, which measure the stability of clocks, such as the plot on slide 4 (and copied below): Allan Deviation plot this level of performance at 1000 seconds is only reached by hydrogen masers. The typical GPS Disciplined Oscillator will be at about 1e-12@1000seconds, see this recent comparison of 17 GPSDOs. With careful correction of the sawtooth error caused by the internal oscillator of the GPS (discussed at the end of the presentation) you can improve the typical 50 ns jitter of GPS to ~5 ns. Beyond this atmospheric/ionospheric effects come in and you need to use a dual-frequency GPS to take those out.

$\endgroup$
1
  • $\begingroup$ You also need phase coherence for times much shorter than 1000sec. Up to the sampling times as well (few nsec). Hydrogen masers don't have very good short term stability so usually crystal oscillator is used for subsecond stability and the signal is 'disciplined' by the maser. Ultimately the maser is needed as a frequency standard rather than for absolute time synchronization. $\endgroup$
    – Chris
    May 28, 2021 at 4:54