I am engineer new to Astronomy and am trying to understand spec of an spectrograph (RV stability = 2m/s). There is a note in document saying "RV shift of 2 m/s is equivalent to a shift of the spectrum on the detector of about 30 nm". The measurement range is 380nm-850nm.

My understanding so far is as follow: Stars moving away/towards earth causes shift in known spectrum lines towards red or blue depending on direction of movement. The magnitude of shift can be derived from Doppler effect as:

$\Delta_\lambda = \lambda_0 \frac{V_r}{c}$

where $\Delta_\lambda$ is the shift in wavelength, $\lambda_0$ is the rest wavelength, and $V_r$ is the radial velocity

or $V_r = c\frac{\Delta_\lambda }{\lambda_{0}}$

RV stability would be difference between radial velocity of consecutive measurements.

Taking an example here:

$V_{r1} = \frac{\Delta\lambda_1 c}{\lambda_0}$ $V_{r2} = \frac{\Delta\lambda_2 c}{\lambda_0}$

where $V_{r1}$ and $V_{r2}$ are the two radial measurements, and $\Delta\lambda_1$ and $\Delta\lambda_2$ are the two wavelength shifts.

$\Delta V_{r} = \frac{(\Delta\lambda_1 - \Delta\lambda_2 )*c}{\lambda_0}$

where $\Delta V_{r} $ is Radial velocity stability between measurements.

$\Delta_s = \frac{\Delta V_r \lambda_0}{c}$

Where $\Delta_s$ is $\Delta\lambda_1- \Delta\lambda_2$ = shift on ccd due to measurement

Plugging numbers from note.

Shift of spectrum on detector = $\frac{2 \cdot 380nm}{3\cdot 10^8} \neq 30nm$

I am definitely doing something wrong but do not know what exactly.

$\endgroup$
1
  • 1
    $\begingroup$ @fasterthanlight Thanks a lot for making it legible. $\endgroup$
    – Nirmala
    Commented Mar 24, 2022 at 22:19

1 Answer 1

1
$\begingroup$

The spectrum is recorded on a detector. The 30nm refers to a physical displacement on the detector.

I e. If you measured the centroid of a spectral line and then applied a 2 m/s shift, the line centroid would move, physically, by 30 nm on the detector.

Your calculation gives the wavelength shift (also in nm) of a spectral line. Combining these calculations tells us that there are roughly 0.084 nm of spectrum per mm on the detector.

Most CCDs have pixels of size $\sim 15$ $\mu$m, so each such pixel would cover 0.00126 nm of wavelength, which is indeed, very high resolution.

$\endgroup$
4
  • $\begingroup$ My understanding is that the $ \Delta_s $ calculated above is the physical shift of spectral lines on detector, which is way off from the document's value. $\endgroup$
    – Nirmala
    Commented Mar 24, 2022 at 23:10
  • 2
    $\begingroup$ @Nirmala no,it isn't You have calculated the shift in wavelength of spectral line. Where in your calculation did you insert the properties of the camera (e.g. the focal length)? $\endgroup$
    – ProfRob
    Commented Mar 24, 2022 at 23:28
  • $\begingroup$ Got it. Thank you for the explanation. $\endgroup$
    – Nirmala
    Commented Mar 24, 2022 at 23:45